Thread: Poiesis Duette Catheter - Balloons Fluid Loss

  1. #1

    Poiesis Duette Catheter - Balloons Fluid Loss

    A while ago, "Scaper1" posted about losing a lot of fluid out of the balloons of the Poiesis Duette Catheters, 10cc and 5cc ports. I decided to experiment with a used catheter and compare 7 days of balloon inflation.

    My Experiment:
    I filled the "non black" (orange in my case-16Fr) port with 10cc of sterile water and the "black" port with 5cc of sterile water, per the instructions for balloon inflation. I placed the catheter on a paper towel which I had labeled with the day of inflation. I took a picture of the inflated balloons. I waited 7 days and took another picture of the same catheter.

    My Results:
    Yes the balloons did lose fluid by a significant amount as you can see in the picture. The fully inflated balloons are clear and stretched out. The balloons, 7 days later are not at clear and visibly smaller. The pictures are taken from the same distance away, or very close to the same distance. I placed the catheter on a paper towel to absorb any liquid. The paper towel was never visibly or to the touch wet. What's more, there was no water mark on the paper towel as there would be if it had been wet and then dried. I presume the fluid slowly evaporated through the silicone membrane of the balloon. After taking the photos, I deflated the balloons. There were a little more than 2cc left in the black port balloon and 7cc left in the non black port balloon. Whether this is the same amount of liquid that would be lost if the catheter was inside the bladder is hard to say.

    Seems like it would be good practice to check the balloons about every 10 days by withdrawing liquid from the balloons and refilling them. I think I would test the black port (cushioning balloon) first. If the balloon had lost 1cc (20%), I think I would just re-inflate the 5cc balloon with 5cc of sterile water, and stop, then possibly repeat the inflation test in a few days. If I found the 5cc balloon had lost more than 20%, I would re-inflate it with 5cc of sterile water and then test the 10cc balloon for water loss.

    I will send my results to Poiesis for their take on this water loss.
    Attached Images Attached Images  
    Last edited by gjnl; 04-06-2017 at 06:32 PM.

  2. #2
    Senior Member
    Join Date
    Feb 2011
    Location
    Washington DC area--------------------T3 complete
    Posts
    115
    Interesting - good experiment !

  3. #3
    Senior Member
    Join Date
    Mar 2005
    Location
    Brooksville, Florida
    Posts
    568
    Quote Originally Posted by gjnl View Post
    A while ago, "Scaper1" posted about losing a lot of fluid out of the balloons of the Poiesis Duette Catheters, 10cc and 5cc ports. I decided to experiment with a used catheter and compare 7 days of balloon inflation.

    My Experiment:
    I filled the "non black" (orange in my case-16Fr) port with 10cc of sterile water and the "black" port with 5cc of sterile water, per the instructions for balloon inflation. I placed the catheter on a paper towel which I had labeled with the day of inflation. I took a picture of the inflated balloons. I waited 7 days and took another picture of the same catheter.

    My Results:
    Yes the balloons did lose fluid by a significant amount as you can see in the picture. The fully inflated balloons are clear and stretched out. The balloons, 7 days later are not at clear and visibly smaller. The pictures are taken from the same distance away, or very close to the same distance. I placed the catheter on a paper towel to absorb any liquid. The paper towel was never visibly or to the touch wet. What's more, there was no water mark on the paper towel as there would be if it had been wet and then dried. I presume the fluid slowly evaporated through the silicone membrane of the balloon. After taking the photos, I deflated the balloons. There were a little more than 2cc left in the black port balloon and 7cc left in the non black port balloon. Whether this is the same amount of liquid that would be lost if the catheter was inside the bladder is hard to say.

    Seems like it would be good practice to check the balloons about every 10 days by withdrawing liquid from the balloons and refilling them. I think I would test the black port (cushioning balloon) first. If the balloon had lost 1cc (20%), I think I would just re-inflate the 5cc balloon with 5cc of sterile water, and stop, then possibly repeat the inflation test in a few days. If I found the 5cc balloon had lost more than 20%, I would re-inflate it with 5cc of sterile water and then test the 10cc balloon for water loss.

    I will send my results to Poiesis for their take on this water loss.
    I check my Duette catheter balloons once a week and usually have to put around 1cc in both balloons.

  4. #4
    Senior Member lynnifer's Avatar
    Join Date
    Aug 2002
    Location
    Windsor ON Canada
    Posts
    18,081
    My normal foley loses about 5-10cc within a month. I fill it to almost 30cc.
    Make America Sane Again. lol

    T-11 Flaccid Paraplegic due to TM July 1985 @ age 12

  5. #5
    This is important to know as I plan on using these catheters next time I need a Foley. If one checks how much fluid is in the balloon and it is insufficient does one use additional prefilled syringes? I believe the Foley syringes are prefilled, correct? Or can one use an empty prefilled to draw in sterile saline from another container to use? Otherwise, one needs a pretty good supply of prefilled syringes on hand.

  6. #6
    You don't need to fill the balloons from prefilled syringes either initially or when you need to add more sterile water.

  7. #7
    I received an email from Poiesis regarding this loss of liquid in the balloons.
    "Osmosis is common with all 100% silicone catheters, it also appears the amount of osmosis occurring varies depending on length of indwelling time as well as the individuals hydration. We conduct tests with each new lot number but usually have the catheters submerged in water since when indwelling they are in a moist to wet environment, having them sitting on a paper towel may have accelerated the loss of fluid. In our testing we have not seen loss of fluid, but have heard from patients with longer indwelling periods."

    It was also suggested that we use pre-filled syringes with sterile water and glycerin solution. This product is sterile and presented in a sterile packet. The glycerin solution has shown to be beneficial in reducing the amount of water lost from the balloon over a length of time, most noticeably in 100% silicone catheters, due to maintaining an osmotic pressure comparable to urine. If you search the internet, you can find sources for these prefilled syringes.
    Last edited by gjnl; 04-08-2017 at 07:31 PM.

  8. #8
    More information about the 10% aqueous glycerin prefilled syringes and why glycerin help to retain liquid in the balloons:
    Attached Images Attached Images  

  9. #9
    They're kindly sending me a sample of the glycerine syringes, so I'll report back too. Currently we're using sterile water, and draining and refilling the balloons every week.

  10. #10
    Senior Member Fragile's Avatar
    Join Date
    Jul 2004
    Location
    Southwest NM
    Posts
    191

    Question Reliable source?

    If you search the internet, you can find sources for these prefilled syringes.
    Has anybody found a reliable source for the prefilled syringes with glycerin?

Similar Threads

  1. Inflating foley catheter balloons
    By Scaper1 in forum Equipment
    Replies: 5
    Last Post: 02-22-2017, 01:26 AM
  2. Duette catheter
    By HACKNSACK44 in forum Care
    Replies: 100
    Last Post: 03-01-2016, 02:35 PM
  3. Break the Balloons
    By 2jazzyjeff in forum Life
    Replies: 1
    Last Post: 01-04-2006, 05:52 AM
  4. Replies: 2
    Last Post: 04-07-2002, 03:37 PM

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •